HP4275A LCR Meter Tutorial
by Paul Birkholz
last updated July 2008
Measure Cµ of a BJT
-
Under Display A,
press
C. Under LCRZ Range,
press Auto. Under
Trigger, press Int.
-
Use the up and
down arrows keys under
Frequency Setup to set the measurement
frequency to 1MHz. The frequency
can be seen in the upper right LED display.
HP4275A front panel.
-
Remove any device from the test fixture.
Press zero open to calibrate the
instrument.
-
Place the transistor in the text fixture as shown in
the figure above. The base should be placed in the side marked
low and the collector should be placed in
the side marked
high. Use the screws to
tighten down the clamps, but do not over tighten.
- A coaxial cable can be found in back of the
instrument that is connected to the DC bias input. Connect this
cable to the nearest power supply (connect + to + and
- to -).
- Set the power supply to the desired value
of VCB.
Read off the value of Cµ
from Display A.
Circuit Models Used in
Calculations
The HP4275A uses the following circuit models and
calculations depending on the configuration of the instrument.
Circuit models and equations for impedance calculations.
|